On May 4, 2009, at 6:48 PM, Maia Cherney wrote:
> The resolution is 2.15 A. The NCS was always on during the refinement
> until we got low R factors (19.2% and 21.2%). Then the NCS was turned
> off for the final refinement and the R factors increased, which is
> strange as they should be going down when you apply less restraints.
This isn't necessarily true - regardless of which program you use to
refine, R-work isn't actually the refinement target, and minimization
algorithms aren't foolproof either. When you remove the restraints
you are effectively decreasing the observation:parameter ratio, which
increases the risk of overfitting. If nothing else I would expect the
gap between Rwork and Rfree to increase, but I've also seen both
increase when NCS restraints were removed.
It may be possible that the NCS restraints are biasing your test set,
leading to the suspiciously low Rfree-Rwork gap, but I thought this
only applied when the NCS approached crystallographic symmetry, and
there's no such thing as fivefold crystallographic symmetry.
Hopefully someone else on the list can clarify this.
-Nat
-------------------