Dear Phenix users, I have a very anisotropic data as phaser reports anisotropic deltaB = 60.2. I would be grateful for advice of several issues. 1.Could you please tell me if phaser map coefficients FWT,PHWT take into account the anisotropic scaling ? 2.This means that these coefficients will be different from those calculated from a partial model in sigmaa because sigmaa has no anisotropic scaling (and no bulk solvent correction) ? 3.In the case of such severe anisotropy can the scaling diminish too strongly the well measured high resolution reflections ? If so should I calculate the coefficients my self by sigmaa and not use pahser mtz output or is there a better solution ? Many thanks in advance for sharing your thoughts and experience, Peter.
Hi, Yes, the data used in the Phaser map coefficients have been rescaled to remove the anisotropy, so the map should look more isotropic than the one you get from running SIGMAA. But note that neither program does a proper bulk solvent correction. Until recently (e.g. in version 2.1.4 of Phaser), the rescaling was done so that the overall average falloff of diffraction was preserved, i.e. the weakest direction was scaled up and the strongest direction was scaled down. However, we were inspired by a paper from Mike Sawaya to look at this again. He showed some convincing results that the maps are more interpretable if the weak data are all scaled up to the falloff of the strongest direction, and the tests we did agreed with this. So that is the behaviour you'll get in recent nightly builds. This agrees with your worry that the scaling could diminish the strongest reflections too much, as happened in the older versions of Phaser. I hope that helps! Regards, Randy Read On 10 Sep 2009, at 12:58, Peter Grey wrote:
Dear Phenix users,
I have a very anisotropic data as phaser reports anisotropic deltaB = 60.2. I would be grateful for advice of several issues. 1.Could you please tell me if phaser map coefficients FWT,PHWT take into account the anisotropic scaling ? 2.This means that these coefficients will be different from those calculated from a partial model in sigmaa because sigmaa has no anisotropic scaling (and no bulk solvent correction) ? 3.In the case of such severe anisotropy can the scaling diminish too strongly the well measured high resolution reflections ? If so should I calculate the coefficients my self by sigmaa and not use pahser mtz output or is there a better solution ?
Many thanks in advance for sharing your thoughts and experience,
Peter.
_______________________________________________ phenixbb mailing list [email protected] http://www.phenix-online.org/mailman/listinfo/phenixbb
------ Randy J. Read Department of Haematology, University of Cambridge Cambridge Institute for Medical Research Tel: + 44 1223 336500 Wellcome Trust/MRC Building Fax: + 44 1223 336827 Hills Road E-mail: [email protected] Cambridge CB2 0XY, U.K. www- structmed.cimr.cam.ac.uk
Dear Prof. Read,
Thank you very much. Will this improved anisotropic scaling be implemented
in other modules of phenix.refine ? If not do you advise using F_ISO,
SIGF_ISO from the newer version of phaser as the input labels for
refinement ?
Peter.
On Thu, Sep 10, 2009 at 2:31 PM, Randy Read
Hi,
Yes, the data used in the Phaser map coefficients have been rescaled to remove the anisotropy, so the map should look more isotropic than the one you get from running SIGMAA. But note that neither program does a proper bulk solvent correction.
Until recently (e.g. in version 2.1.4 of Phaser), the rescaling was done so that the overall average falloff of diffraction was preserved, i.e. the weakest direction was scaled up and the strongest direction was scaled down. However, we were inspired by a paper from Mike Sawaya to look at this again. He showed some convincing results that the maps are more interpretable if the weak data are all scaled up to the falloff of the strongest direction, and the tests we did agreed with this. So that is the behaviour you'll get in recent nightly builds. This agrees with your worry that the scaling could diminish the strongest reflections too much, as happened in the older versions of Phaser.
I hope that helps!
Regards,
Randy Read
On 10 Sep 2009, at 12:58, Peter Grey wrote:
Dear Phenix users,
I have a very anisotropic data as phaser reports anisotropic deltaB = 60.2. I would be grateful for advice of several issues. 1.Could you please tell me if phaser map coefficients FWT,PHWT take into account the anisotropic scaling ? 2.This means that these coefficients will be different from those calculated from a partial model in sigmaa because sigmaa has no anisotropic scaling (and no bulk solvent correction) ? 3.In the case of such severe anisotropy can the scaling diminish too strongly the well measured high resolution reflections ? If so should I calculate the coefficients my self by sigmaa and not use pahser mtz output or is there a better solution ?
Many thanks in advance for sharing your thoughts and experience,
Peter.
_______________________________________________ phenixbb mailing list [email protected] http://www.phenix-online.org/mailman/listinfo/phenixbb
------ Randy J. Read Department of Haematology, University of Cambridge Cambridge Institute for Medical Research Tel: + 44 1223 336500 Wellcome Trust/MRC Building Fax: + 44 1223 336827 Hills Road E-mail: [email protected] Cambridge CB2 0XY, U.K. www- structmed.cimr.cam.ac.uk
_______________________________________________ phenixbb mailing list [email protected] http://www.phenix-online.org/mailman/listinfo/phenixbb
Hi, I've been talking to Pavel about whether or not phenix.refine should use the same approach to the map coefficients. He's always busy implementing a lot of new things, so I'm not sure if he's had a chance to evaluate it for himself. At the moment, if you want this behaviour you'll have to use F_ISO, SIGF_ISO from recent versions of Phaser. Ultimately, of course, it would be better if you used the original data in phenix.refine and it made the desired map coefficients after carrying out its own correction. Regards, Randy On 10 Sep 2009, at 17:18, Peter Grey wrote:
Dear Prof. Read,
Thank you very much. Will this improved anisotropic scaling be implemented in other modules of phenix.refine ? If not do you advise using F_ISO, SIGF_ISO from the newer version of phaser as the input labels for refinement ?
Peter.
On Thu, Sep 10, 2009 at 2:31 PM, Randy Read
wrote: Hi, Yes, the data used in the Phaser map coefficients have been rescaled to remove the anisotropy, so the map should look more isotropic than the one you get from running SIGMAA. But note that neither program does a proper bulk solvent correction.
Until recently (e.g. in version 2.1.4 of Phaser), the rescaling was done so that the overall average falloff of diffraction was preserved, i.e. the weakest direction was scaled up and the strongest direction was scaled down. However, we were inspired by a paper from Mike Sawaya to look at this again. He showed some convincing results that the maps are more interpretable if the weak data are all scaled up to the falloff of the strongest direction, and the tests we did agreed with this. So that is the behaviour you'll get in recent nightly builds. This agrees with your worry that the scaling could diminish the strongest reflections too much, as happened in the older versions of Phaser.
I hope that helps!
Regards,
Randy Read
On 10 Sep 2009, at 12:58, Peter Grey wrote:
Dear Phenix users,
I have a very anisotropic data as phaser reports anisotropic deltaB = 60.2. I would be grateful for advice of several issues. 1.Could you please tell me if phaser map coefficients FWT,PHWT take into account the anisotropic scaling ? 2.This means that these coefficients will be different from those calculated from a partial model in sigmaa because sigmaa has no anisotropic scaling (and no bulk solvent correction) ? 3.In the case of such severe anisotropy can the scaling diminish too strongly the well measured high resolution reflections ? If so should I calculate the coefficients my self by sigmaa and not use pahser mtz output or is there a better solution ?
Many thanks in advance for sharing your thoughts and experience,
Peter.
_______________________________________________ phenixbb mailing list [email protected] http://www.phenix-online.org/mailman/listinfo/phenixbb
------ Randy J. Read Department of Haematology, University of Cambridge Cambridge Institute for Medical Research Tel: + 44 1223 336500 Wellcome Trust/MRC Building Fax: + 44 1223 336827 Hills Road E-mail: [email protected] Cambridge CB2 0XY, U.K. www- structmed.cimr.cam.ac.uk
_______________________________________________ phenixbb mailing list [email protected] http://www.phenix-online.org/mailman/listinfo/phenixbb
_______________________________________________ phenixbb mailing list [email protected] http://www.phenix-online.org/mailman/listinfo/phenixbb
------ Randy J. Read Department of Haematology, University of Cambridge Cambridge Institute for Medical Research Tel: + 44 1223 336500 Wellcome Trust/MRC Building Fax: + 44 1223 336827 Hills Road E-mail: [email protected] Cambridge CB2 0XY, U.K. www- structmed.cimr.cam.ac.uk
Sharpening the map to aid the human eye is one thing, modifying the data you put into refinement is another matter completely. People have been sharpening low resolution maps for model-building for years, anisotropy has nothing to do with it. It seems to be easier for the eye to pick out features in the contours of a sharpened map when the resolution is low. There is no reason to expect that the sharpness of the data in the best diffracting direction is the optimal sharpness for model-building. You may wish to sharpen more depending on your data set and your eyes. Refinement programs don't care. If you up weight the high resolution terms your average B will drop, the Fcalc's also up weight and you end up with the same difference coefficients, the same shifts, and the same results. All that happens of significance is that you have inappropriately lowered your atomic B factors and your Table 1 gives the reader an overoptimistic view of the quality of your model. If you have a low resolution map, by all means go ahead and look at a sharpened version (as I said, anisotropy has nothing to do with it). You must, however, refine your model against the data you actually collected. Dale Tronrud Peter Grey wrote:
Dear Prof. Read,
Thank you very much. Will this improved anisotropic scaling be implemented in other modules of phenix.refine ? If not do you advise using F_ISO, SIGF_ISO from the newer version of phaser as the input labels for refinement ?
Peter.
On Thu, Sep 10, 2009 at 2:31 PM, Randy Read
mailto:[email protected]> wrote: Hi,
Yes, the data used in the Phaser map coefficients have been rescaled to remove the anisotropy, so the map should look more isotropic than the one you get from running SIGMAA. But note that neither program does a proper bulk solvent correction.
Until recently (e.g. in version 2.1.4 of Phaser), the rescaling was done so that the overall average falloff of diffraction was preserved, i.e. the weakest direction was scaled up and the strongest direction was scaled down. However, we were inspired by a paper from Mike Sawaya to look at this again. He showed some convincing results that the maps are more interpretable if the weak data are all scaled up to the falloff of the strongest direction, and the tests we did agreed with this. So that is the behaviour you'll get in recent nightly builds. This agrees with your worry that the scaling could diminish the strongest reflections too much, as happened in the older versions of Phaser.
I hope that helps!
Regards,
Randy Read
On 10 Sep 2009, at 12:58, Peter Grey wrote:
> Dear Phenix users, > > I have a very anisotropic data as phaser reports anisotropic deltaB > = 60.2. I would be grateful for advice of several issues. > 1.Could you please tell me if phaser map coefficients FWT,PHWT take > into account the anisotropic scaling ? > 2.This means that these coefficients will be different from those > calculated from a partial model in sigmaa because sigmaa has no > anisotropic scaling (and no bulk solvent correction) ? > 3.In the case of such severe anisotropy can the scaling diminish too > strongly the well measured high resolution reflections ? If so > should I calculate the coefficients my self by sigmaa and not use > pahser mtz output or is there a better solution ? > > Many thanks in advance for sharing your thoughts and experience, > > Peter. > > _______________________________________________ > phenixbb mailing list > [email protected] mailto:[email protected] > http://www.phenix-online.org/mailman/listinfo/phenixbb
------ Randy J. Read Department of Haematology, University of Cambridge Cambridge Institute for Medical Research Tel: + 44 1223 336500 Wellcome Trust/MRC Building Fax: + 44 1223 336827 Hills Road E-mail: [email protected] mailto:[email protected] Cambridge CB2 0XY, U.K. www- structmed.cimr.cam.ac.uk http://structmed.cimr.cam.ac.uk
_______________________________________________ phenixbb mailing list [email protected] mailto:[email protected] http://www.phenix-online.org/mailman/listinfo/phenixbb
------------------------------------------------------------------------
_______________________________________________ phenixbb mailing list [email protected] http://www.phenix-online.org/mailman/listinfo/phenixbb
Hi Peter,
Will this improved anisotropic scaling be implemented in other modules of phenix.refine ? If not do you advise using F_ISO, SIGF_ISO from the newer version of phaser as the input labels for refinement ?
1) phenix.refine does anisotropic correction, see this paper for more details: Acta Cryst. (2005). D61, 850-855. "A robust bulk-solvent correction and anisotropic scaling procedure". P. V. Afonine, R. W. Grosse-Kunstleve and P. D. Adams. and slide #20 here: http://phenix-online.org/presentations/neutron_japan_2009/phenix_japan_part1... 2) Improved version of anisotropic scaling suggested by R. Read (more precisely, a better way of map calculation given high anisotropy) will be implemented in future. We discussed it but I didn't get a chance yet to implement it. 3) I do not believe that it is a good idea to use manipulated Fobs in refinement. Pavel.